Arterial pressure in the radial artery is 155/70 mmHg measured by a correctly calibrated catheter-transducer system. At the same time, aortic pressure is 140/75 mmHg using a high-fidelity catheter tip transducer. The most likely cause of this discrepancy is
(A) a large amount of air in the dome of the radial artery transducer
(B) coarctation of the aorta
(c) peripheral vascular constriction produced by sympathetic stimulation
(D) physiologic amplification of the waveform from the aorta to the radial artery
(E) too high a frequency response in the catheter-transducer system
A 27-month-old, 14-kg infant with intestinal obstruction is anesthetized with nitrous oxide at 1 L/min, oxygen at 0.4 L/min, and halothane at 0.8% using a Bain circuit. Ventilation is controlled at a rate of 30/min. The child's temperature is 39°C, and the PaCO2 is 65 mmHg. Which of the following is most likely?
While checking an anesthesia machine, opening the oxygen flow-control valve yields no oxygen flow, although the wall-mounted oxygen pipeline supply gauge reads 50 psig. Opening the backup oxygen cylinder results in normal oxygen flow. The most likely cause is
(A) failure of the oxygen pipeline supply
(B) failure of the second-stage oxygen pressure regulator
(C) a malfunctioning check valve in the oxygen pipeline supply inlet
A healthy patient is receiving general endotracheal anesthesia with nitrous oxide 4 L/min, oxygen 2 L/min, and isoflurane 1% with spontaneous ventilation. The inspiratory breathing hose becomes disconnected from the inspiratory valve. The earliest alarm will be produced by
(A) a pulse oximeter with finger probe set to alarm at 60% saturation
(B) a mass spectrometer set to alarm when no breath is detected for 30 seconds
(C) an oxygen analyzer mounted on the inspiratory valve outlet and set to alarm at 25% oxygen concentration
(D) a capnograph sensor at the Y-piece that alarms when end-tidal carbon dioxide tension is greater than 60
Two opened oxygen cylinders, one full and one empty, are attached to an anesthesia machine. The check valve to the empty cylinder fails. This is expected to result in
(A) activation of the fail-safe device
(B) cessation of oxygen flow to the anesthesia circuit
(C) filling of the empty cylinder with a mixture of all gases supplied to the anesthesia machine
(D) increase in the temperature of the empty cylinder
(E) interruption of gas delivery from the central oxygen source
A 30-year-old man who is undergoing laparotomy and resection of a large kidney tumor has a decrease in SpO2 from 100% to 92% and an increase in peak airway pressure from 20 to 35 cm H2O. Plateau pressure is unchanged at 18 cm H2O. Which of the following is the most likely cause?
An otherwise healthy 20-year-old man undergoes laser surgery for vocal cord polyp removal. During jet ventilation with 100% oxygen, arterial blood gas analysis shows PaO2 90 mmHg, PaCO2 40 mmHg, and pH 7.40. Which of the following is the most likely explanation?
During anesthesia with nitrous oxide, oxygen, and a relaxant, the patient's lungs are being mechanically ventilated and there is a heated humidifier in the circuit. The most likely cause of an apparent increase in inspired oxygen concentration is
(A) a hole in the ventilator bellows
(B) backflow of fresh gas because of resistance in the humidifier
An air/oxygen blender with E-size cylinders is supplying gas to a Mapleson D circuit being used to ventilate a neonate during transport. The air tank is full (psi= 1900/625 L) and flow is 4 L/min. The oxygen tank flow is 5 L/min and the pressure gauge indicates 500 psi. If the flow rates remain unchanged, which of the following statements is correct?
(A) The air tank will be empty in 90 minutes
(B) The oxygen tank will be empty in 30 minutes
(C) The concentration of oxygen in the circuit is 60%
(D) The oxygen tank will fill with air when the pressure gauge falls below 50 psi
(E) The volume of oxygen remaining can be determined only by weighing the tank
A healthy, 18-year-old, 65-kg woman undergoing diagnostic laparoscopy is being monitored with a properly functioning and calibrated capnograph. The end-tidal carbon dioxide concentration is 6%, and the inspired carbon dioxide concentration is 1%. This may be caused by each of the following EXCEPT
(A) malfunction of the expiratory valve in a circle system
(B) low gas inflow with a Bain circuit
(C) exhausted soda lime within a circle system
(D) absorption of carbon dioxide from carbon dioxide laparoscopy
During use of the ventilator on an anesthesia machine, positive pressure is noted on the airway pressure gauge during exhalation. Positive end-expiratory pressure has not been purposefully added to the breathing circuit. Which of the following is the most likely cause?
(A) Closure of the pop-off valve in the circle system
(B) Excessive tidal volume settings on the ventilator
(C) Obstruction of the pressure relief valve on the scavenging system
(D) Overinflation of the endotracheal tube balloon
A patient whose trachea is intubated develops partial airway obstruction two hours after induction of fentanyl, nitrous oxide, oxygen anesthesia. The anesthesiologist discontinues the nitrous oxide and controls ventilation, and there is gradual improvement during the next hour. No evidence of a respiratory disorder is found at the conclusion of the anesthetic. Which of the following is most likely?
(A) The endotracheal tube was in the right mainstem bronchus
(B) The patient had an unrecognized air embolism
(C) The patient had a nitrous oxide pneumothorax
(D) The endotracheal tube cuff narrowed the airway
During induction of anesthesia, a patient is breathing isoflurane from a semiclosed circle absorber system. The temperature-compensated vaporizer is set to deliver isoflurane 1.5%. Which of the following will result in the lowest inspired concentration of isoflurane?
(A) Doubling the length of circle system hoses
(B) Decreasing the barometric pressure
(C) Changing from rubber to plastic breathing hoses
Following antagonism of a nondepolarizing muscle relaxant with neostigmine, twitch height is normal. To achieve this effect, the percentage of the acetylcholine receptor pool that must be free of muscle relaxant is
Postoperatively, a patient is being mechanically ventilated by a constant-flow, pressure-cycled ventilator with the following initial settings: inspiratory/expiratory (I/E) ratio of 1:2, peak inspiratory pressure (PIP) of 25 cmH2O, and rate of 10/min. One hour later, the I/E ratio is 1:4. Which of the following would ensure that the minute ventilation is the same as that initially set?
(A) Inflate the endotracheal tube cuff to prevent leakage
(B) Double the respiratory rate
(C) Decrease the expiratory pause until the I/E ratio is 1.0
During attempted cannulation of the right internal jugular vein, the exploring needle goes lateral to the vein. Which of the following structures is most likely to be encountered?
A Fluotec vaporizer is set to administer halothane 4% using a fresh gas flow of 4 L/min. If the diluent gas flow is increased to 8 L/min, and at the same time the room temperature increases from 20 to 25°C, the approximate concentration of halothane delivered to the patient would be
An isoflurane-specific vaporizer that is 25% full is mistakenly refilled with halothane, and the dial is set at 1.5%. Which of the following statements concerning this situation is true?
(A) The gas mixture will contain equal partial pressures of halothane and isoflurane
(B) The gas mixture will produce a greater depth of anesthesia than 1.5% isoflurane alone
(C) The gas mixture will produce a greater depth of anesthesia than 1.5% halothane alone
(D) The liquid mixture will have a higher vapor pressure than 1.5% halothane alone
(E) The liquid mixture will have a higher vapor pressure if it is placed in a halothane-specific vaporizer
An 88-kg man is anesthetized with 66% nitrous oxide, 33% oxygen, and 1% isoflurane using a circle system. Following tracheal intubation, the nitrous oxide flow is set at 1 L/min and the oxygen flow is set at 0.5 L/min. After two hours the oxygen monitor indicates a circuit oxygen concentration of only 25% despite the flows noted above. The most likely explanation for this discrepancy is
(A) decrease of oxygen pipeline pressure to 20 psi
(B) increase in the concentration of nitrous oxide in the circuit
A 68-year-old patient is undergoing total hip replacement during general anesthesia using positive pressure ventilation and neuromuscular blockade. Intraoperative monitoring shows a gradual decrease in end-tidal CO2, an increase in exhaled tidal volume, and an increase in measured FiO2. Which of the following is the most likely cause of these changes?
(A) Air entrainment around a deflated endotracheal tube cuff
(B) Channeling of the exhaled air through the carbon dioxide canister
(C) Fat embolism
(D) Inadequate muscle relaxation with spontaneous patient ventilation
A 16-year-old child has an arterial catheter inserted prior to a potentially difficult intubation for a craniofacial operation. While awake and breathing room air, arterial blood gas values are: PaO2 134 mm Hg, PaCO2 28 mm Hg, and pH 7.45. These values can be explained by
(A) excess heparin mixed with the arterial sample
(B) a patient core temperature of 38°C
(C) excess air entrained in the syringe during sampling
(D) storing the sample at room temperature for 15 minutes before processing
Which Mapleson breathing circuit shown in the illustration permits the LEAST amount of fresh gas inflow to prevent rebreathing during spontaneous ventilation?
Arterial pressure in the radial artery is 155/70 mmHg measured by a correctly calibrated catheter-transducer system. At the same time aortic pressure is 140/75 mmHg using a high-fidelity catheter tip transducer. The most likely cause of this discrepancy is
(A) a large amount of air in the dome of the radial artery transducer
(B) coarctation of the aorta
(C) peripheral vascular constriction produced by sympathetic stimulation
(D) physiologic amplification of the wave form from the aorta to the radial artery
(E) too high a frequency response in the catheter-transducer system
A 30-kg patient with asthma is receiving general anesthesia and being mechanically ventilated with a measured tidal volume of 300 mL. Increasing the fresh gas flow from 3 L/min to 9 L/min without changing the ventilatory rate or I:E ratio will result in
A normal adult is receiving general anesthesia via a standard circle system and controlled ventilation. Mass spectrometer data are shown. Which event is most compatible with these data?
A radial artery catheter is to be used for blood pressure measurement during a sitting craniotomy. When zeroing the transducer, which of the following describes the best levels for placement of the transducer and opening of the system to air? (Transducer, Opening to Air)
After inflation of both the tracheal and endobronchial cuffs of a left-sided double-lumen tube, the following are noted on auscultation of the chest: Left lumen clamped --> Breath sounds absent bilaterally; Right lumen clamped --> Breath sounds present on left only; Both lumens open --> Breath sounds present on left only. The most appropriate action is to
(A) decrease the endobronchial cuff volume
(B) decrease the tracheal cuff volume
(C) reinsert the tube, rotating it 180 degrees to the right
(D) suction both lumens
(E) withdraw the tube slightly and repeat the auscultory maneuvers
The two E oxygen cylinders on an anesthesia machine have pressure readings of 1100 psi each. At an oxygen flow of 3 L/min, there will be sufficient oxygen for approximately
During laser microsurgery of the larynx using an endotracheal tube, a fire occurs in the airway. Which of the following is the most appropriate initial management?
The lungs of a 7-kg infant are being ventilated with a volume-cycled ventilator at a rate of 20/min. The anesthetic circuit has a compressible volume of 5 ml/cm H2O, and the peak inspiratory pressure is 20 cmH2O. To achieve adequate ventilation, the ventilator tidal volume should be set at
During general anesthesia using a modern anesthesia machine, a patient's spontaneous ventilation is manually assisted with a fresh gas flow of 5 L/min. A sudden malfunction of the pressure-relief (pop-off) valve of the circle absorption system occurs, and the valve cannot be opened. The most appropriate next step is to
If the low-pressure alarm of an anesthesia circuit is broken, which of the following monitors will provide the earliest indication of a disconnected ventilator hose?
Oxygen 100 ml/min is bubbled through a vaporizer containing an anesthetic with a vapor pressure of 150 mmHg, and this mixture is added to a fresh gas flow of 5 L/min. The delivered anesthetic concentration is
A 35-kg child requires mechanical ventilation with pure oxygen at a tidal volume of 350 ml and a rate of 20/min during a severe asthma attack. The most likely cause of severe hypotension after initiating mechanical ventilation is
During general anesthesia administered through a circle system, the soda lime absorbant is exhausted. No fresh soda lime is available for use. Which of the following is the most appropriate next step to prevent hypercapnia in this patient?
(A) Decreasing the dead space of the circle system
The capnographic tracing is from a 2-month-old infant anesthetized using a pediatric circle system and mask at a fresh gas flow of 4 L/min. The sampling port is in the elbow connector. This tracing indicates
(A) adequate alveolar ventilation
(B) exhausted soda lime
(C) expired halothane concentrations representative of alveolar concentrations
An unconscious adult patient is being ventilated through an esophageal obturator airway (EOA) in the emergency department. In the absence of cervical spine injury, which of the following is appropriate?
(A) Use of the EOA for airway management until the patient regains consciousness
(B) Removal of the EOA before insertion of an endotracheal tube
(C) Placement of an endotracheal tube before removal of the EOA
(D) Removal of the EOA under fiberoptic endoscopic visualization
A 55-year-old man is undergoing craniotomy in the sitting position. Mean arterial pressure is 75 mmHg; arterial blood gas values are PaCO2 41 mmHg and pH 7.37. End-tidal CO2 is 7 mmHg. Which of the following is the most likely cause of the increased PaCO2 to PetCO2 gradient?
(A) Decreased cardiac output
(B) Endobronchial intubation
(C) Hyperinflation of the lungs
(D) Partial disconnect of the capnograph sample tubing
The two E oxygen cylinders on an anesthesia machine have pressure readings of 1100 psi each. At an oxygen flow of 3 L/min, there will be sufficient oxygen for approximately
A waste-gas scavenger has an active disposal system. Which of the following safety features is necessary to prevent removal of excess gas from the breathing circuit?
An anesthesia machine is set to deliver oxygen 2 L/min, nitrous oxide 2 L/min, and sevoflurane. After 30 minutes of stable anesthesia, which of the following, is the most likely cause of a decrease in the oxygen analyzer reading from 50% to 30%?
(A) A leak in the ventilator bellows
(B) Accumulation of water on the oxygen sensor membrane
(C) Calibration at 100% oxygen using less than 100% oxygen
(D) Disconnection of the wall oxygen hose
(E) Presence of the oxygen analyzer in the expiratory limb